Hi there
Is it possible to have two devices connected at the same time to a CSI/O port in a CR1000 logger?
In this case we want to communicate two loggers via RF416, but at the same time we need to have a MD485 module connected to the CSI/O port, is this ok?
Thanks
The configuration you describe should work fine as long as the RF416 and the MD485 are assigned different SDC address settings. I would suggest using the network planner to generate the settings.
That´s what I though,
thanks a lot
This video: https://www.campbellsci.com/videos/loggernet-software-network-planner shows how to use the Network Planner.