I recently updated the LoggerNet to ver 4.5 and found out that it is not possible to update the constant table of my CR1000 in the Connect Window. I've done this before under LoggerNet ver 4.4.2 and it worked. Wondering what's wrong on it.
Do you see the table name in the software?
Does the datalogger have OS 28 or later? If not, then the Table Name you give for ConstTable will not show up in the software.
Hi Terri
The OS is ver 31, and I can see the ConstTable on software. I tried to update the value and changed the "ApplyAndRestart" into true or 1, but the value just kept turning into the orginal one.
I am not having trouble with Loggernet 4.5 and CR1000.Std.31 using Connect Screen to make the change.
What are you useing to set ApplyAndRestart?
I have used Table Monitor and Numeric Display which are both in the Connect screen.
Are you still having problems?
If so, be sure to press the Enter key after you change each value. If you don't press Enter, the values aren't really being set, including ApplyAndRestart.