Not getting good data from my Setra 278 on a new CR300


TedDarby Oct 15, 2018 05:22 PM

Had to swap out a CR200x with a CR300 this morning after the 200's serial port gave up. Everything was working fine yesterday up until then. I had Campbell update our program a couple of weeks ago for the new 300's but when I programed it and hooked up the 278 it keeps giving dumb data like the unit is not connected in the data fields.

Currently the CR300 is reporting:

BP_mb = 640.6462
BP_inHg = 18.91828
BP_psi = 9.289371

Last night the readings on the CR200 before it died were

BP_mb = 989.0448
BP_inHg = 29.20649
BP_psi = 14.34115

I am under the impression that from what I have seen and what I have read that the connections have change a bit between CR200 and CR300

Currently I have the Setra wired as follows:

Setra 278 wiring block:
Ext Trigger: Green
AGND: Yellow
GND: Black
Supply: Red
Vout: Blue

CR300:
Green: C1
Yellow: G
Black: G
Red: 12v +
Blue: SE

Any suggestions?

I have a weather station right next to this setup and it is reporting 30.11 inHg at this moment.


JDavis Oct 17, 2018 07:27 PM

If you are reading 640 mB, that is near the offset. So, that would be about 0 volts. Make sure your program is turning on C1 to power up the sensor.