oabany Mar 18, 2023 03:48 AM

Hello community,

I have a CR1000 with two sensors:
HMP45C
CS616

I used shortcut to generate the program and when using pc400 in the public table the values for CS616 are always 0.00

I have double checked the wiring diagram and I don't know why it is 0.00.

do I need to calibrate it?

how can I check it is able to work and it is not broken?


Thanks so much, any help is much appreciated.


JDavis Mar 20, 2023 06:54 PM

Look at the value for the period. That is the one with the default lable of PA_uS.

The period is the raw reading before any calibration is appied.