AVW200 frequency range


TLT Aug 12, 2024 02:41 PM

Hi Everyone,

I have a question about AVW200 frequency range. Is it fine for me to use a widen frequency sweep range than the sensor frequency range. I want to fit two different frequency range sensor (Geokon model 4000: 450Hz to 1250Hz and Geokon model 4450: 1200Hz to 2800Hz) to one same multiplexer. 

I set the begin and end frequency to 450 and 2800 to fit both sensors.

 AVW200 (AVW_Result(1),ComC1,0,200,DST(1,1),1,1,32,450,2800,2,_60Hz,1,0) 

The logger was able to read the sensors and the Hz readings were very close compare to the readings taken with a vibrating wire indicator. My concerns are does it affect the accuracy of the readings? 

Please advise. Thanks.

Regards,

TLT


JDavis Aug 12, 2024 07:28 PM

You are fine using the wider frequency range.

The reason for narrowing a frequency range would be to eliminate a known source of noise.