ljsmall44 Mar 21, 2022 07:22 PM

Hi all,

I am working on a program for structural testing that is using a Volt 116 for linear potentiometers and a VWire 305 for vibrating wire guages. I am having trouble figuring out the calibration for both the Volt 116 and VWire 305 when coding a program for both of them to run together with the VWire using a 20hz mainscan and the Volt 116 using a 1hz SlowSequence scan. 

When in shortcut, I can get the potentiometers and load cell to calibrate well, however, in the combined program, it is not working so I was hoping I could rebuild the program in shortcut, but I cannot find the Granite VWire 305 device in the shortcut devices menu. Am I missing something here, or is it just no compatible?


nsw Apr 6, 2022 03:48 PM

The VWire 305 units are not supported in ShortCut.

You will need to develop this in CRBasic. There are some example programs that you can download from here :-

GRANITE VWIRE 305: 8-Channel Dynamic Vibrating-Wire Analyzer (campbellsci.co.uk)

Log in or register to post/reply in the forum.